In Admin
 Sensor Configuration, you can view the Sensor Health column to monitor sensor health and you can open a sensor's chart to see its health trend. This allows you to check if a sensor is receiving data as expected.
To view sensor health status:
-  In Admin
 Sensor Configuration, make sure the Sensor Health column is enabled in the table.
		 - If it is not, click on Columns and enable it.
 -  You can also enable the Last Reading Date/Time column to see when the last reading was received
To view sensor health trends:
 Use the three dots menu (
 Chart to view the sensor's uptime for the last 1 hour, 8 hours, 24 hours, and 7 days. You can also view a historical chart with the sensor's data points.
Sensor health calculations
Uptime calculations are based on the interval set for each sensor and the data points received in the period (last 1 hour, 8 hours, etc.).
The sensor health status specifically is based on the uptime for the last 24 hours.
Sensor health status is defined as follows:
- Good = > 94.9% Uptime
 - Fair = Uptime between 84.9% and 94.9%
 - Poor = < 84.9 % Uptime
				   (Count # of received data points in last 24 hours) / ((86400 seconds in 24 hours) / (Sensor interval)) * 100  |
For example, for a sensor with a sampling interval of 900 seconds and 82 received data points in the last 24 hours:
				 82 / (86400 / 900) * 100 = 85%  |